I am going to estimate retention index on HP-5 ms UI column (30 x 0.25 x 0.25). The potential problem is I have only mixture hydrocarbons C10-C40. I am afraid that would be a problem with elution long-chained compounds. Does anybody has an expierience with it?

Re: Kovats index on HP-5ms

Posted: Fri Jun 26, 2015 8:05 pm
by Peter Apps
Welcome to the forum.

I am not sure that I understand your problem - do you have components of your sample that have more than 40 carbon atoms, or do you anticipate problems eluting the heavier molecules in your alkane standard mixture ?

Peter

Re: Kovats index on HP-5ms

Posted: Sun Jun 28, 2015 6:45 am
by Robert Zakrzewski
Hello Peter,
I mean the second problem, I usually analyse flavours and worry that with my mixture (C10-C40) on that column I could observe heavy hydrocarbons for many injections in my following anayses.

Re: Kovats index on HP-5ms

Posted: Sun Jun 28, 2015 10:33 am
by Peter Apps
I see. A 40-carbon alkane will take a long time to elute. If you cannot source a standard without the heavier hydrocarbons than you may need to run your retention standard as the last run of the day and leave the column at high temperature overnight. It might also be a good idea to increase the inlet temperature, and the split ratio while the column is baking out.

Peter

Re: Kovats index on HP-5ms

Posted: Tue Jun 30, 2015 7:29 pm
by Robert Zakrzewski
Thank you Peter,
I found application for HP-5ms for hydrocarbons up to C28, but I think I can try your suggestion for my mixture.