Strange pressure problems
Posted: Fri Jun 26, 2015 9:47 am
Hi there - I didn't find anything similar to this, so I'll open up a new discussion.
We're running an Agilent 1260 System with a quat pump (G1311B) and noticed strange pressure drops last week when my colleague made some clean-up runs (we're using it semiprep most of the time).
The flow rate will always show 1mL/min and the pressure stays stable at 200bar and suddenly drops to ~1-3 bar, and will then ripple from there to 30 bars and will eventually go up again. We measured the actual flow and it is as expected less than 1mL/min.
I then exchanged the frit and gold seal (and later on also the whole purge valve, which isn't very old btw) and it seemed that the problem has been solved. At least for ~10 runs and the problem occured again, even with the new PV. I took a look under the microscope with our "old" PV but didn't notice anything broken.
I took a look at the pistons (seemed good I think..) and exchanged all the seals in the pump head. I also connected water directly to the inlet valve (passive inlet valve) and put a restriction capillary to the other end to exclude a problem with the column. Still nothing.
The Agilent technician was here 3 times now and just left and left the pump running at 1mL/min with 50/50 MeOH/H2O and this is what the pressure-meter is telling me:
The picture attached continues with a short time period of 70 bar, after that 10 minutes of rippling 0-30bar, then again it goes up to 70 bar for 10 mins and then 200 bar.
He also exchanged the MCGV and degassing unit, also the inlet and outlet valves; he also changed the eluent-inlet-filters and also tried it without any filters.
Did anyone ever had this kind of problem? I'd be glad to hear some cheering words before we must buy a new pump or something