i am using FID with db5 cap column.
nitrogen carrier gas, hydrogen and air.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Hi all. Too lazy to translate myself, komputer-iron, let it workthedude,
You are looking at a test plot rather than the detector signal. In the signal part of the method you should be looking at sig1 or sig2 for the active detector and not test plot.
Best regards,
AICMM
If this was solvent and impurity peaks it would not keep repeating through the run. AICMM is right - this is the HP Agilent test signal.Hi all. Too lazy to translate myself, komputer-iron, let it workthedude,
You are looking at a test plot rather than the detector signal. In the signal part of the method you should be looking at sig1 or sig2 for the active detector and not test plot.
Best regards,
AICMM![]()
![]()
![]()
Perhaps worse, if a person has never worked with a chromatograph.
He sees a large peak of the solvent and impurities to the solvent
A desired peak can not see because there are not looking. (in first min)
1) Use a slow gradient from 100 C. The first minutes of cut off - there is a solvent. Cannabidiol fly at 200 C.
2) Carefully annealing column (325С reverse gas flow - revese column ) to remove the dirt in the column.
3) Put more of cannabis, to see exactly rush.
I meant it.i bought an agilent 5890 series II GC for testing medical cannabis samples. when i run the machine with any method, i get i large peak followed by one medium peak followed by one very small hump. this keep continually repeating every minute as long as a method is running. I am totally new to gc, so am at somewhat of a loss. can anyone give me some idea what might be causing this problem and what i can do to correct it. i just want to get the machine running with a good baseline. not worried about testing ATM.
i am using FID with db5 cap column. nitrogen carrier gas, hydrogen and air.
that was the problem. thanks.You are looking at a test plot rather than the detector signal. In the signal part of the method you should be looking at sig1 or sig2 for the active detector and not test plot.
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.