Problems with the TOGA method!
Posted: Wed Jun 24, 2015 3:39 pm
We recently acquired a new Agilent 7820A gas chromatographer (adapted with an Agilent G4367A Auxiliary Electronic Pressure Control (Aux EPC) Module) with a 7697A Headspace, with the following specification (according to ASTM D3612 Method C):
• Carrier: Argon
• Detectors:
o thermal conductivity detector (TCD)
o flame ionization detector (FID)
• Columns:
o PLOT 5 A Mol. Sieve Column
o Carboxen-1006 PLOT Column
We verify our gas supply to be 99.999% pure, Nitrogen, Hydrogen, Argon and Air.
Using the standard gas references (100 ppm, 1000 ppm and 5000 ppm) we perform regular calibrations in order to obtain accurate results.
Using the ASTM D3612 method C we have set up the correction equation to obtain the corrected results at 760 torr and 273 Kelvin. We use a multiplier for the data acquisition of 1/Lab temperature (°K).
PROBLEM: We have recently performed an interlaboratory and found that our results vary significantly. Mostly in our H2, N2, CO2, CO and O2 results with differences of over 70%. I have checked all the formulas and the method itself and its all as it should be, however our results do not fall within the 2 standard deviation.
I would like any opinions/ suggestions as to what I can do to correct this.
Thank you,
V.H
• Carrier: Argon
• Detectors:
• Columns:
Thank you,
V.H