syx » Tue Sep 06, 2005 1:47 am
Depends on exactly what you mean by "could not be separated".
What was the resolution between the peaks on the "good" instrument, versus on the "bad" instrument?
On “normal condition” we found a peak when medium dissolution (water) was injected. So, we called it the ‘peak of water’ – Peak A. When we injected folic acid solution in medium dissolution, we found 2 peaks: Peak A and Peak B (folic acid). On LC A, we could find these 2 peaks, but not in LC B.
Was the loss in resolution caused by:
- a significant shift in retention both peaks?
- a significant shift in retention of one of the peaks?
- excessively wide or tailing peaks?
- complete disappearance of one peak?
In LC B, there is significant shift in retention of Peak B and slight shift of Peak A compared to LC A. The peaks in LC A are in good shape: sharp and without tail. The appearance in LC B is shouldering peak in front slope of Peak B. When we injected medium dissolution, we found that Peak A was retained right on front slope of Peak B. So, the shouldering shape on Peak B was Peak A.
Are you premixing your mobile phase or are you mixing "on-line"?
If premixing, have you tried using the same batch of mobile phase on both instruments?
We use “on-line” mixing.
Note: Before the LC run for this research, the tube between column outlet and detector inlet was cracked and a hole was created on the part near the column outlet (+/- 3 cm from the column outlet). I cut the tube and then fixed it again to the column outlet. Could this length reduction be as the reason of the incident?