I have Spectrasystem TSP, which is equipped with a UV2000 detector. Unfortunately I can not connect this instrument to a computer because I do not have RS232 to DB9 adapter. Would someone be so kind who has a similar system to display scheme for connecting pins inside the adapter.
-
- Posts: 36
- Joined: Wed Jul 31, 2013 9:29 am
-
- Posts: 18
- Joined: Sat Dec 14, 2013 11:40 am
Thank you very much for the website, but it helped me a lot when someone gave the scheme for connecting pins inside the adapter. This would mean scheme of RS232 (9 or 25 pins) and DB9 (8 pins) and appropriately connecting wires between these two connectors. Of course, it would be the best if someone has this instrument configuration and to see how the wires are connected inside the adapter.

-
- Posts: 327
- Joined: Sun Jan 01, 2012 10:21 am
As I understand , you need the connection diagram of the adapter connected to RS232 port of the computer , which has a 8 pin DB9 input.
9 pin RS232 8 pin DB9
1 No connection
2 6
3 5
4 1
5 3
6 8
7 7
8 2
9 No connection
The pin numbers stated as seen from the front side of the sockets.
I hope it helps
9 pin RS232 8 pin DB9
1 No connection
2 6
3 5
4 1
5 3
6 8
7 7
8 2
9 No connection
The pin numbers stated as seen from the front side of the sockets.
I hope it helps
-
- Posts: 18
- Joined: Sat Dec 14, 2013 11:40 am
Thank you very much Uzman.
I just have one little question, whether the scheme of pins connecting from the adapter of this Thermo HPLC instrument? I ask this because on the web you can find a huge number of schemes between these two ports.
Thanks again.
I just have one little question, whether the scheme of pins connecting from the adapter of this Thermo HPLC instrument? I ask this because on the web you can find a huge number of schemes between these two ports.

Thanks again.
-
- Posts: 327
- Joined: Sun Jan 01, 2012 10:21 am
Hello Emmet ,
This is exactly from my own Thermo , P1000 pump , AS3000 autosampler and UV2000 detector.
This is exactly from my own Thermo , P1000 pump , AS3000 autosampler and UV2000 detector.
-
- Posts: 18
- Joined: Sat Dec 14, 2013 11:40 am
Thanks a lot once again, uzman.

-
- Posts: 18
- Joined: Sat Dec 14, 2013 11:40 am
Dear Uzman,
Did you under the designation DB9 thought the socket for LAN cable? If you did, have you included the pinout (pin numbers) as given at these links:
http://www.tomsitpro.com/articles/local ... 263-3.html
The upper left image below following text (socket for LAN cable):In the following image, an 8P8C (RJ45) cable connector is wired to the AT&T 258A/EIA 568B standard.. (pin number for LAN socket)
and
http://www.scourge.fr/mathdesc/lb/lb-8.html
The first image from the top (pin number for rs 232).
The important thing is just to confirm that you're in this way numbered pins. Everything else on these links is not important. Also, if there is something other than a LAN socket and rs 232, between the computer and the SN4000 controller, it would be good if you can describe me.
Thanks in advance.
Did you under the designation DB9 thought the socket for LAN cable? If you did, have you included the pinout (pin numbers) as given at these links:
http://www.tomsitpro.com/articles/local ... 263-3.html
The upper left image below following text (socket for LAN cable):In the following image, an 8P8C (RJ45) cable connector is wired to the AT&T 258A/EIA 568B standard.. (pin number for LAN socket)
and
http://www.scourge.fr/mathdesc/lb/lb-8.html
The first image from the top (pin number for rs 232).
The important thing is just to confirm that you're in this way numbered pins. Everything else on these links is not important. Also, if there is something other than a LAN socket and rs 232, between the computer and the SN4000 controller, it would be good if you can describe me.
Thanks in advance.
-
- Posts: 327
- Joined: Sun Jan 01, 2012 10:21 am
I uploaded the connection diagram to the following link :
http://i62.tinypic.com/34y8tch.jpg
Please be aware that , the socket pins stated as seen from the front ,for both sockets.
I also numbered the 8P8C( RJ45) socket myself , from left to right ( 1 to 8 ) .
On the web , it is numbered from right to left , please be careful about this.
9 pin RS232 socket is correctly numbered.
http://i62.tinypic.com/34y8tch.jpg
Please be aware that , the socket pins stated as seen from the front ,for both sockets.
I also numbered the 8P8C( RJ45) socket myself , from left to right ( 1 to 8 ) .
On the web , it is numbered from right to left , please be careful about this.
9 pin RS232 socket is correctly numbered.
-
- Posts: 18
- Joined: Sat Dec 14, 2013 11:40 am
Dear, Uzman
Thank you very much for your efforts and help.
Thank you very much for your efforts and help.

-
- Posts: 18
- Joined: Sat Dec 14, 2013 11:40 am
Dear Uzman,
Looks like I made a good adapter, but in the meantime I found that I do not have another part, and that's another problem for which the instrument probably does not work. Part name is a contact closure wiring harness (A5596-010). Could you draw a diagram inside the black box, if it is not too complicated. From what I understand it's a box of contacts with wires. If it's not too complicated I'd try to make one by myself.
Thank you in advance.
PS: Sorry, maybe I'm boring but I really want to run that old HPLC instrument.
Looks like I made a good adapter, but in the meantime I found that I do not have another part, and that's another problem for which the instrument probably does not work. Part name is a contact closure wiring harness (A5596-010). Could you draw a diagram inside the black box, if it is not too complicated. From what I understand it's a box of contacts with wires. If it's not too complicated I'd try to make one by myself.
Thank you in advance.
PS: Sorry, maybe I'm boring but I really want to run that old HPLC instrument.

-
- Posts: 327
- Joined: Sun Jan 01, 2012 10:21 am
Hi Emmet,
There is no black box on my system but there are some wire connections behind the instruments.
I think this is what you need.
Look at the back side of the instruments , there are some terminals.
Detector has 12 pin , Autosampler has 12 pin and Pump has 8 pin terminals.
Connect the pin 10 of detector to pin 6 of autosampler and pin 7 of pump.
Those pins must be connected together.
Also connect pin 9 of detector to pin 1 of autosampler and pin 3 of pump.
This is ground connection.
What operating system do you use , OS/2 or Windows ?
Some TSP systems use OS/2 based software and they don't need the above wiring.
If you use Windows based ThermoQuest software ( version 4.0 or above ) you must use the wiring.
If you are trying to convert a OS/2 based system to Windows based ThermoQuest software , you must also change the EPROMs in the SN4000 controller , otherwise it will not work.
There is no black box on my system but there are some wire connections behind the instruments.
I think this is what you need.
Look at the back side of the instruments , there are some terminals.
Detector has 12 pin , Autosampler has 12 pin and Pump has 8 pin terminals.
Connect the pin 10 of detector to pin 6 of autosampler and pin 7 of pump.
Those pins must be connected together.
Also connect pin 9 of detector to pin 1 of autosampler and pin 3 of pump.
This is ground connection.
What operating system do you use , OS/2 or Windows ?
Some TSP systems use OS/2 based software and they don't need the above wiring.
If you use Windows based ThermoQuest software ( version 4.0 or above ) you must use the wiring.
If you are trying to convert a OS/2 based system to Windows based ThermoQuest software , you must also change the EPROMs in the SN4000 controller , otherwise it will not work.
-
- Posts: 18
- Joined: Sat Dec 14, 2013 11:40 am
Hello Uzman,
I am using Windows NT 4.0, because I read in some guide I found on the net. Regarding the software, I also found some software called Xcalibur 1.2. The following links are pieces of software that I used during the complete installation:
http://sjsupport.thermofinnigan.com/dow ... =Go&id=402
http://sjsupport.thermofinnigan.com/dow ... =Go&id=403
Also my SN4000 controller contains the following:
In the socket U12 is located a chip that reads: (V 4.69 A4636-127), in the socket U29 is located a chip that reads: (V 4.69 A4657-297). Sockets U11 and U28 are empty.
What else can I try in order to run instrument?
Thank you in advance.
I am using Windows NT 4.0, because I read in some guide I found on the net. Regarding the software, I also found some software called Xcalibur 1.2. The following links are pieces of software that I used during the complete installation:
http://sjsupport.thermofinnigan.com/dow ... =Go&id=402
http://sjsupport.thermofinnigan.com/dow ... =Go&id=403
Also my SN4000 controller contains the following:
In the socket U12 is located a chip that reads: (V 4.69 A4636-127), in the socket U29 is located a chip that reads: (V 4.69 A4657-297). Sockets U11 and U28 are empty.
What else can I try in order to run instrument?
Thank you in advance.
-
- Posts: 327
- Joined: Sun Jan 01, 2012 10:21 am
Hello Emmet ,
You can not run your software with these chips on Windows , they are suitable for OS/2 .
You need V.9.05-A4636-129 chip for running under Windows.
Only one chip is necessary for socket U12 , other sockets will be empty for Windows.
You can not run your software with these chips on Windows , they are suitable for OS/2 .
You need V.9.05-A4636-129 chip for running under Windows.
Only one chip is necessary for socket U12 , other sockets will be empty for Windows.
-
- Posts: 18
- Joined: Sat Dec 14, 2013 11:40 am
Hello Uzman,
Can you tell me what is the name of software that is compatible with OS/2 operating system, and where to be found? Or at least where I can find the right chip for Windows?
Thanks in advance!
Can you tell me what is the name of software that is compatible with OS/2 operating system, and where to be found? Or at least where I can find the right chip for Windows?
Thanks in advance!

32 posts
Page 1 of 3
Who is online
In total there are 58 users online :: 0 registered, 0 hidden and 58 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 58 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 58 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science