Inconsistent Nalorphine
Posted: Thu Jun 11, 2015 1:15 pm
				
				In our tox lab we analyze blood and urine for morphine, codeine, hydrocodone, hydromorphone, 6-MAM, oxycodone, and oxymorphone as TMS derivatives with nalorphine as an internal standard (qualitative detection only, no quants). We use SPE and GCMS. We recently cut the nalorphine concentration in urine down due to trace morphine showing up in the negative controls, and now occasionally the nalorphine drops out almost completely or at least to failing level. It is currently 500 ng/mL (50 uL spiking volume) using a 1 mL sample of urine.
The problem happens randomly regardless of analyst, PPM, turbovap, etc. and happens to controls as well as samples. We adjust the pH to 6 prior to adding to the SPE column. We have collected the sample load, methanol wash, and water wash and see no breakthrough. An extra aliquot of elution solvent (MeCl/IPA/NH4OH 78/20/2) shows no extra being retained after elution. We have been unable to force a failure by playing with flow rates. In controls we will see a large drop in nalorphine with little to no drop in other drugs, so it seems to be specific to nalorphine. It is very sporadic and the drop amount is inconsistent. We are being extra careful about spiking the IS. Can anyone shed some light on what might be happening? We are working on a new method with deuterated IS but need to fix this in the meantime. I can share more specifics about the method if needed.
			The problem happens randomly regardless of analyst, PPM, turbovap, etc. and happens to controls as well as samples. We adjust the pH to 6 prior to adding to the SPE column. We have collected the sample load, methanol wash, and water wash and see no breakthrough. An extra aliquot of elution solvent (MeCl/IPA/NH4OH 78/20/2) shows no extra being retained after elution. We have been unable to force a failure by playing with flow rates. In controls we will see a large drop in nalorphine with little to no drop in other drugs, so it seems to be specific to nalorphine. It is very sporadic and the drop amount is inconsistent. We are being extra careful about spiking the IS. Can anyone shed some light on what might be happening? We are working on a new method with deuterated IS but need to fix this in the meantime. I can share more specifics about the method if needed.