				high purge flow measured on P&T
				Posted: Tue Jun 09, 2015 7:03 am
				by tangaloomaflyer
				We have a system that has very high purge flow (120ml/min) when measured through the vent of the concentrator.  Visually, the bubbles do not appear to be forming at a much higher rate than an instrument purging at 40ml/min.  Am I right in assuming there must be a constriction in the flow path of the concentrator?  Samples are being purged in an Archon autosampler and the sample gas is connected to the T-piece on a Velocity concentrator.  the system is having problems holding a calibration.  It will be steady fro a while but then responses will fluctuate.
			 
			
					
				Re: high purge flow measured on P&T
				Posted: Tue Jun 09, 2015 7:19 am
				by kubowicz.tomasz
				Hello
What kind of glassware are you using for purging? Frit or fritless ?
Tomasz Kubowicz
			 
			
					
				Re: high purge flow measured on P&T
				Posted: Wed Jun 10, 2015 2:33 am
				by tangaloomaflyer
				Fritless.  10ml of aqueous sample is in a gas tight vial.  It is purged by the soil probe on the autosampler.  No heat, 7 minute purge.
			 
			
					
				Re: high purge flow measured on P&T
				Posted: Mon Jun 15, 2015 5:50 pm
				by Yama001
				Perhaps both the Archon and the Velocity are purging? The purge supply line to the sparger is diverted to a solenoid valve on the left rear of the Archon - in water mode it simply diverts the flow back to the sparger; in soil mode it is blocked, so only the Archon soil pruge line reaches the concentrator.
			 
			
					
				Re: high purge flow measured on P&T
				Posted: Tue Jun 16, 2015 7:20 am
				by tangaloomaflyer
				I will double check, but we are running the Archon in soil mode.