linearity problem
Posted: Tue Jun 02, 2015 2:19 pm
Hello guys,
I need a help regarding validation-linearity. I have to validate the assay of alpha tocopherol and phytosterol. Y intercept of Tocopherol is very bad: >6%, but the result of phytosterol is good (2,4%).
Do you have any idea?
Here is the method details below:
oven:300°c, He - Injector:320°c - detector (FID): 320°c- Splitflow: 52ml/min.
sampling parameters:
FillStrokes = 3
AirVolume = 2.0 [µl]
FillVolume = 1.0 [µl]
DownSpeed = 20 [mm/s]
SampleDepth = Bottom
DrawSpeed = NotViscousDefault
PullupDelay = 3.0 [s]
DelayAfterBubbleElimination = 2.0 [s]
BubbleEliminationSpeed = 80.0 [µl/s]
DelayBetweenStrokes = 3.0 [s]
Linearity Tocopherol:
Conc.(mg/ml) % Average area RF
1,6556 68,98 0,251924 0,152165
1,9009 79,2 0,295744 0,155581
2,1372 89,05 0,332373 0,155518
2,3743 98,93 0,367273 0,154687
2,5806 107,53 0,406002 0,157329
2,8384 118,27 0,450041 0,158554
3,0794 128,31 0,489726 0,159033
corr. coefficient: 1,53 %
corr. coefficient: 0,99965794; intercept: -0,02297248; x variable: 0,00398958 and the y intercept of 100%: -6,2%(Spez: <+-3%).
Thanks.
I need a help regarding validation-linearity. I have to validate the assay of alpha tocopherol and phytosterol. Y intercept of Tocopherol is very bad: >6%, but the result of phytosterol is good (2,4%).
Do you have any idea?
Here is the method details below:
oven:300°c, He - Injector:320°c - detector (FID): 320°c- Splitflow: 52ml/min.
sampling parameters:
FillStrokes = 3
AirVolume = 2.0 [µl]
FillVolume = 1.0 [µl]
DownSpeed = 20 [mm/s]
SampleDepth = Bottom
DrawSpeed = NotViscousDefault
PullupDelay = 3.0 [s]
DelayAfterBubbleElimination = 2.0 [s]
BubbleEliminationSpeed = 80.0 [µl/s]
DelayBetweenStrokes = 3.0 [s]
Linearity Tocopherol:
Conc.(mg/ml) % Average area RF
1,6556 68,98 0,251924 0,152165
1,9009 79,2 0,295744 0,155581
2,1372 89,05 0,332373 0,155518
2,3743 98,93 0,367273 0,154687
2,5806 107,53 0,406002 0,157329
2,8384 118,27 0,450041 0,158554
3,0794 128,31 0,489726 0,159033
corr. coefficient: 1,53 %
corr. coefficient: 0,99965794; intercept: -0,02297248; x variable: 0,00398958 and the y intercept of 100%: -6,2%(Spez: <+-3%).
Thanks.