UltiMate 3000 WPS-3000 autosampler needle replacement proble

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hello all,

I recently had reset my autosampler to factory settings, changing the puncture depth and my needle broke upon injection. I'm now having injection problems which I hope to fix by inquiring if I made a mistake upon replacing the needle my self.

I was hoping someone had one of the 2.4 uL replacement syringes handy and could provide me with some measurements. Mainly, the total length and the distance from the tip to the rubber piece where it bends for the holder.

I used a micrometer to measure the OD as 280 um and based on a length estimate of our old needle ( which I have no idea if it was Dionex/Thermo's or silica a previous lab member replaced ) of ~ 28 cm the ID should be 100 um.


Upon breaking and my own measurements, I replaced the needle with ~ 28 cm of 300 OD 100 ID um silica. I changed the puncture setting to the max of 11 mm, used the Chromeleon command to puncture our sample vials, fed the silica down to the bottom, and after hitting the bottom pulled up a little. Much like packing fused silica in the high pressure bomb.

My Flush 1 and overfill should be sufficient for the needle volume and loop. A difference measurement of a sample vial with water showed that the needle did pick up some liquid on "inject" and the needle does release liquid into the wash port when I use that command. Upon injection of sample I'm not seeing the expected results. I'm looking into my sample prep, but that hasn't changed and I'm not even seeing the unretained junk in the chromatogram.

I'm considering the draw velocity but I don't want to start playing with that until I'm positive the problem isn't before that.

( Wow that was more than I was planning to write. )

Much obliged.
I hope it helps. Sorry for low quality.

Disclaimer: Currently, I am not using this capillary. It lies in a drawer and I truly hope that this is a working version.

Image


Image


Image
HPLC 2017 in Prague, http://hplc2017-prague.org/
ID is 100 um (checked by SEM)
HPLC 2017 in Prague, http://hplc2017-prague.org/
Jiri,

I had estimated around 12 cm based on my described method.

The quality is perfect for the measurement. I am extremely grateful!

Much obliged! :D
4 posts Page 1 of 1

Who is online

In total there are 36 users online :: 1 registered, 0 hidden and 35 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 35 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry