severe broadening in gcms

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
I am running a 5973/6890 as a new install. I ran a linear gradient from 40 to 300 over 54 minutes. about halfway thru
the run the peaks on the MS side start broadening badly and after 30 minutes come out as one wide hump.
I have never run without the thermal aux to the MS side. Is this due to adsorption in the transfer to the MS?
Can a thermal aux be added?

Thanks!
Marc


[/img]Image
Define what top and bottom chromatograms are.

What is your MSD temperature?
Thanks CPG. Top is MS and bottom is FID. Running post-column split. Quad is 150 and source is 230.

I never connected the aux heater cable. Vented and connected and now working fine.
FYI the aux setting is what heats the transfer line. Before you connected it you were condensing sample after the GC oven. Did you cut the portion of the column that was in the transfer line when you vented?
4 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry