Regular noise increase

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hello
Maybe someone will give an idea what is the source of this strange noise increase
Multiresidue method LC-MS/MS (Agilent 6410A), ESI+
Example sequence (samples):
1/2/3/4
5 - noise increase
6/7
8 - noise increase
9/10
11 - noise increase
12/13
14 - noise increase
15/16
17 - noise increase
...
- noise increase occurence independent on matrix (the same phenomena for samples/calibrants/reagent blanks etc.)
- area/hight for analyte signal stable
- noise increase like 2x
http://snag.gy/UbRdD.jpg

Printscreen of TIC for sample 38 (black line,noise increase) and reinjection 38A (red line,normal signal)
http://snag.gy/qAYkg.jpg
A. You mention that this is a multi-residue LC-MS/MS analysis, and show the noise for the m/z 327 to 77 transition.
1) is there a similar behavior for all transitions ?

B. It is very obvious that the behavior occurs at every 3rd injection, AFTER the first 5 injections in the example shown
1) is this behavior observed for every sample set ?
2) Does your analysis sequence go,
Blank/ low level to high level calibn stds/Blank/ QC Sample/ Blank/Sample 1/Sample 2 etc. ?
3) is the behavior observed if blanks are regularly inserted in the sequence ?

C. The appearance of the TIC suggests that the m/z 327 analyte is bleeding slowly into the source, rather than being a distinct chromatographic peak. This suggests a possible hang-up of the analyte in the LC lines and/or ion source. Is there any chance that this is column bleed from a Phenyl-type column.
1) does your HPLC method include a high-% organic wash at end of individual analysis ?
2) what column are you using ?
3) what is mobile phase ?

JMB
Hello JMB,
Thanks for reply.
Yes its MRM method. Noise increase is similar across all transitions. The 327->77 is just for example that peak area is stable (its my ISTD - TPP).
Every 3rd injection after the first increase of signal (its random from 1 to 6...trying to find the rule of it)
Sequence: reagent blank/low level to high level calibrants/QC sample/blank/samples 1...i dont insert any blanks in sequence just calibrant on LOQ level.
I wash HPLC/nebulizer with high% organic (9:1 ACN/H2O and the end of each batch).
Column: Zorbax Eclipse Plus 100x2.1 1.8um
Mobile phase:
A: H2O 0.01% Formic acid, 5mM Ammonium formate
B: 9.5:0.5 ACN/H2O 0.01% Formic acid, 5mM Ammonium formate

I elongate organic stage in gradient ... and set few (10) no injection at the end of sequence will check results tomorrow.
I wash HPLC/nebulizer with high% organic (9:1 ACN/H2O and the end of each batch).

Is yours a gradient HPLC method, and does it include,

a) a high-% organic wash after each INJECTION ?
and
b) a needle wash/rinse ?

JMB
4 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry