Demonstration of specificity
Posted: Wed May 20, 2015 9:53 am
by jayroseville58
If impurity or degradation product standards are unavailable, specificity may be demonstrated by comparing the test results of samples containing impurities or degradation products to a second well-characterized procedure (e.g., a pharmacopeial or other validated procedure). These comparisons should include samples stored under relevant stress conditions (e.g., light, heat, humidity, acid or base hydrolysis, oxidation)
I'm confuse with the above mentioned statement. Need some clarification. Thanks.
Re: Demonstration of specificity
Posted: Fri May 22, 2015 11:46 am
by danko
Which part of it confuses you?
Posted: Tue May 26, 2015 1:03 pm
by jayroseville58
is forced degradation studies should be always applied on our RS test method validation? thanks.
Posted: Tue May 26, 2015 1:15 pm
by danko
Yes.
Unless you already know in detail the degradation products and have them available so that you can demonstrate adequate separation. The latter (the separation og the degradants from the intact molecule) and determination of those is the specificity indicator here.
Posted: Thu Jun 04, 2015 7:38 am
by jayroseville58
Yes.
you mean to say, if we are able to show specificity of impurities with the use of its corresponding available standard, it is no longer required to us to conduct forced degradation?
Posted: Thu Jun 04, 2015 8:13 am
by danko
It is my conception anyway. Just make sure you have the right arguments for the impurities you’re focusing on. That is to say they are relevant to the drug you’re dealing with, i.e. that is what will happen upon formulation and storing the product under relevant conditions.