Method development
Posted: Mon May 18, 2015 11:21 am
by brasileo_88
I was interessted in determining the assay and related substances of spectinomycin by HPLC. I know about a method that uses ECD detector.
I intented on using other method because i haven't got the ECD detector. The method consist of a normal phase and using a derivatization agent. Does any one know if this method is useful?
The normal phase method using UV Detection was taked from Chromatography journal 333.
brasileo@gmail.com
Re: Method development
Posted: Mon May 18, 2015 12:04 pm
by Fernando
Hi brasileo_88
Consult this article.
A Validated Stability-Indicating HPLC Method for Routine Analysis of an Injectable Lincomycin and Spectinomycin Formulation
Murad N. Abualhasan,*,1 Nidal Batrawi,2 Oliver B. Sutcliffe,3 and Abdel Naser Zaid1
Best luck
Fernando
Posted: Mon May 18, 2015 1:55 pm
by brasileo_88
Thanks Fernando
I'll try
Posted: Tue May 19, 2015 1:35 pm
by artsjeroen
I don't know what is in your sample matrix, but if the selectivity of ECD is not necessary I would go for HPLC UV.
We measure the activity of Spectinomycin using reversed phase on C18 with UV detection (215 nm).
Using SDS in the mobile phase for ion pairing. Works well for us, with good symmetrical peaks.
Posted: Tue May 19, 2015 3:46 pm
by brasileo_88
Thanks for your answer.
It is listen good, may you have operating condition for you method?
It is valid for related substances too?
Posted: Wed May 20, 2015 10:00 am
by artsjeroen
60% - 2 g SDS + 15 ml H3PO4 in 1000 ml water (pH = 6.0 with Ammonia).
40% - Acetonitrile
flow = 1 ml/min
T = 35°C
Column is symmetry C18, 5 µm, 3.9 x 150 mm, but I guess any C18 with comparable dimensions will work.
Our solutions are the raw material dissolved in water (ca. 2.5 mg/ml).
We don't measure related substances but you could give it a try.
Please let me know if it works for you.
Posted: Wed May 27, 2015 7:45 am
by brasileo_88
Thanks but your method can not be use to deremined related subst.
Posted: Thu May 28, 2015 7:06 am
by artsjeroen
Did it work for assay determination? What was the retention time of Spectinomycin? And how was the peak shape?
With altering column length/mobile phase I think you can get more resolution on the chromatography of the related substances, I don't know why it would not work.
