If the sample contains no volatiles I do not get this issue, with duplicate injections giving very similar values.
I have performed 5 injections from the same well with samples containing the same MeOH content as my assay samples and I do not see a steady decrease in peak area with each subsequent injection, just a difference between the first injection and any of the subsequent injections
I only load 7 ul and the repeat injection is always performed within a maximum of 30 min (minimising any issues associated with sample evaporation) so I'm happy that the issue is not due to depletion of the sample between injections.
Ideally I would like to pierce the film, withdraw the needle, wait a while THEN take the first injection. Unfortunately the closest I think I can come to this is to perform a zero volume injection but considering the excruciatingly slow pace of the needle carriage (this is an old Alliance HT 2795!!) this would mean increasing my run time from around 20 hrs to 30 hrs per assay, significantly affecting my throughput.
As I'm looking at changes of [test article] as measured by ratio to the internal reference I can actually extract the results I want with a high level of consistency and accuracy, I just hate seeing such wide error bars.
Sooooooo..........Has anyone else observed this before, have any ideas what might be causing this variation, and perhaps have a solution which wouldn't increase my run times by up to 1/3.
Thank you