~100mAU blocks in UV-Vis on Dionex Ultimate 3000
Posted: Thu May 14, 2015 9:39 am
Hi all,
I am currently experiencing an issue with big square blocks of signal appearing in my chromatograms. They are always ~100mAU and there is no disturbance or anything out of the ordinary with the pump pressure at the same point in the run. The blocks themselves are raggedy, not clean, so I think it must be a detector issue.
Things I have tried:
I have tried multiple columns, eluents and buffers and the issue occurs sporadically with all of them.
I have replaced the UV-Vis lamps with our back-up and that has not resolved the issue.
I have purged each solvent bottle twice before running.
I have left the system equilibrating for ~2 hours, still hasn't resolved it.
I think it may be a bubble in the detector that is occasionally oscillating in front of the lamp, causing a signal disturbance for 1-3mins, then moving back out, this makes sense as the pressure does not fluctuate. If this is the problem, what is the best way to wash it out? Would disconnecting the column and washing it through continuously be the best plan of action?
Thanks a lot!
I am currently experiencing an issue with big square blocks of signal appearing in my chromatograms. They are always ~100mAU and there is no disturbance or anything out of the ordinary with the pump pressure at the same point in the run. The blocks themselves are raggedy, not clean, so I think it must be a detector issue.
Things I have tried:
I have tried multiple columns, eluents and buffers and the issue occurs sporadically with all of them.
I have replaced the UV-Vis lamps with our back-up and that has not resolved the issue.
I have purged each solvent bottle twice before running.
I have left the system equilibrating for ~2 hours, still hasn't resolved it.
I think it may be a bubble in the detector that is occasionally oscillating in front of the lamp, causing a signal disturbance for 1-3mins, then moving back out, this makes sense as the pressure does not fluctuate. If this is the problem, what is the best way to wash it out? Would disconnecting the column and washing it through continuously be the best plan of action?
Thanks a lot!