Paraquat/Diquat on HILIC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hello,

we currently try to set up a method for Paraquat/Diquat with LC-MSMS on a zwitterionic HILIC column but have trouble to get them eluted from the column. We use 20mMol Ammoniumformate pH=3 and Acetonitril as mobile phases. We tried columns from different suppliers (Sequant ZicHILIC and MN Nucleodur HILIC) but we have to lower the percentage of ACN down to 20% to get them properly eluted from the columns. From the available applications they should be eluted much easier.

HPLC-System: Agilent 1290

Has anyone observed the same problem and any idea why we observe this phenomenon and why and where stick on the column?

Best
Axel
There is a validated method for paraquat and diquat on our Obelisc R mixed-mode column which is currently running in about 80 organizations worldwide.

Here is a reference manual (validated in Europe): http://www.crl-pesticides.eu/library/do ... _QuPPe.pdf
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
I think you might want something less polar for your stationary phase if you're having such a hard time. I'd also suggest a milder pH (4-4.5) for your buffer.

I've used Waters' Amide BEH to separate polysaccharides, and it worked beautifully. Being that you've got a charged analyte, I'm not sure if its the right choice, but its a start. A Zwitterionic column is overkill polarity for your analytes, and may not work at all to do what you want.

The first problem I can see is that the stationary phase is charged, as are your analytes. The analytes are almost definitely adhering to your column electrostatically, and unless you're using a seriously salty buffer (read: instrument damage) you're not going break that interaction terribly easy without a high percentage of water.

Do you have a 1290 binary or 1290 quaternary pump?
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 3 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Google [Bot], John Guajardo and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry