RID problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

9 posts Page 1 of 1
Hey guys.

I´m having trouble with a RID USP technique.
Conditions:
MP: Acetonitrile:Water (79:21)
Flow 2 mL/min
Teperature between 20 - 25 +/- 2 ºC
Column: L8 250 x 4.6 5 um
Working in an Agilent 1100
The problem I have is when I pump through the column. Lot of noise. Pressure is stable, temperature too. Left stabilizing all night and the problem persists. When you put a conector instead the column, noise dissapears, and does the same when turned pump of. No drift.
I´ve tried different columns, but the problem is the same. Tried to clean the cell, and passed Isopropanol because I suspected bubble prescence at first, but nothing.
What could be happening? Any useful advice?
Q. F. Ignacio Viera
How do you degass your mobile phase?
On the outlet of your RID please connect a short capillary, lets say 20 - 30cm max but check the back pressure, that should avoid that on a column you will get micro bubbles and in the detector cell. Good luck.
Gerhard Kratz, Kratz_Gerhard@web.de
Degasser on-line.
Q. F. Ignacio Viera
Degasser on-line.
Hi,
I think that first solution - special degasser procedure of solvents, before hplc experiment. Please use ultrasonic bath or other procedures for this. Second recommendation - only isocratic mode, from the one bottle.

Thank you
When the path is over and passed successfully, it's nice to remember even the mistakes ...
Already did that. And I´m using only channel A.
Q. F. Ignacio Viera
Hello

Do not check pressure/baseline with zero volume connector (it doesn't generate sufficient backpressure so it will tell you nothing). Try restriction capillary instead or small diameter capillary (backpressure at least 100bar).
I would also check:
-capillary between column and detector - perhaps it is too wide so try one with smaller diameter.
-pump valves (leak test)
-optical unit/RI cell temperature (set it to 40deg)

Best regards

Tomasz Kubowicz
Second recommendation - only isocratic mode, from the one bottle.
Already did that. And I´m using only channel A.

FYI - always purge ALL channels even if you only are using one or two.

Also, when we do RID or conductivity detector, even though isocratic, we always bypass the 1100 MCGV (multi-channel gradient valve) to get a smoother baseline. Agilent has a little adapter part for that, included with a new instrument, looks much like a tan fingertight column nut (Agilent Technologies #0100-1847).
I´m by-passing the multi gradient.
Q. F. Ignacio Viera
Route the waste and recycle tubing to an upper position , around half a meter, for Agilent RID.
This is recommended by Agilent.
9 posts Page 1 of 1

Who is online

In total there are 6 users online :: 2 registered, 0 hidden and 4 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 4 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry