Helium Detection Limit by GC-TCD

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
I am wondering what is the typical Helium detection limit by GC-TCD. The method I am using: Agilent 7890GC with a TCD. Column: packed molsieve 13x column; carrier gas: Ar; Oven temp: isothermal @30°C; flow rate: 30ml/min. TCD temp: 200°C. Sampling size: 5ml.

With this method, I am able to achieve a LOQ approximate 5ppm, is this a reasonable limit for He or do I have some room to improve the sensitivity? I am use to seeing ppb level of other gases (such as Ar, N2, CH4 etc) analyzed by GC-PDHID.

Looking forward to hearing your comments. Thank you!
Low ppm for any gases is good sensitivity for TCD as you have seen an HID is much more sensitive
You might (heavy emphasis on the might) be able to get it down to 1-2 ppm, but 5 ppm is not bad. Try balancing your measuring/reference cell flows; this might help a little. You might get a slightly sharper peak with nitrogen carrier; make sure you're running at optimum flow.

TCD just isn't a very sensitive tool, unfortunately.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry