We use an OI 4660 concentrator with an OI 4551 autosampler and an Agilent 6890/5973. The column is a RTX-624, 0.25mm ID, 1.4 um film, 30 m length. The inlet is set for a 40:1 split, column flow is 1 ml/min. The temperature program is 1 min at 40 degrees, 15 degrees per min to 220, total run is 20 min.
The P&T is set to purge 11 min, dry purge 3 min, Desorb 1.5 min at 240 degrees, bake 7 min at 250. Desorb flow is 40 ml/min, using a VOCARB trap.
When we use this for EPA 8260 work, we routine have trouble with the first two surrogates, Dibromofluoroethane and 1,2-Dichloroethane-d4. The associated internal standard is Fluorobenzene.
We get acceptable initial calibrations, but as samples run, the Fluorobenzene internal standard recovery drops, driving up the two surrogate recoveries. Note that we do not use an internal standard spiking module; we hand spike everything. Note also that the rest of our internal standards and surrogates behave well. The system passes all the leak checks. We get acceptable data when the 4600 is used with a 4552 (Archon) autosampler.
I suspect the hand spiking of the internal/surrogates has something to do with it, but I cannot see where.
Thanks, Tim














 
																							 .
.