Residual Solvents by Headspace - Repeatability issue
Posted: Tue May 12, 2015 8:44 am
by ChemStation1
Has anyone experienced high %RSD when performing residual solvents analyes by headspace using headspace autosampler?
I'm using water as diluent for tert-Butanol and using the parameters specified in USP 467.
Any suggestions for critical parameters that may affect Repeatability?
Re: Residual Solvents by Headspace - Repeatability issue
Posted: Tue May 12, 2015 8:50 am
by AaronAIT
Might be an obvious one, but - how do you tighten your HS caps?
Posted: Tue May 12, 2015 9:04 am
by ChemStation1
Using a standard manual crimper.
Is there a significance for using 10ml or 20ml HS vials?
Posted: Tue May 12, 2015 9:16 am
by ChemStation1
I should also mention that the area od tert-Butanol is decreasing after each injection
Posted: Tue May 12, 2015 9:34 am
by AaronAIT
If all HS caps are crimped tight, not dented, and you can't turn them by hand, that should be tight enough.
Perhaps the problem is related to your HS sampler? How does repeatability look for other analytes?
Posted: Tue May 12, 2015 9:38 am
by Peter Apps
What septa are you using ?
Posted: Tue May 12, 2015 10:01 am
by ChemStation1
As for the repeatability of other analytes, Acetonitrile gives the same RSD as tert-butanol.
The septa is thermo scientific- Silicone/PTFE
Posted: Tue May 12, 2015 10:07 am
by AaronAIT
Have you performed any diagnostics on your HS sampler, such as a vent valve test? Your service engineer should be able to guide you through this procedure, it's fairly easy and quick.
Posted: Tue May 12, 2015 11:33 am
by ChemStation1
Thank you.
Do you know what are the recommended settings for the headspace sampler:
Fill speed
pullup delay
injection speed
pore inject delay
post inject delay
Thank you.
Posted: Tue May 12, 2015 2:32 pm
by Bigbear
To make informed suggestions we need your GC and headspace parameters. Also what do you define as high % RSD?
Posted: Wed May 13, 2015 4:49 am
by ChemStation1
Headspace CTC Conditions
Incubation Temp.: 80 C
Incubation Time: 40 min
Syringe Temp: 90 C
Fill Speed-500ul/s
pullup delay-1000ms
pre inj delay-500ms
post inj delay-500ms
Flush time-200sec
GC Conditions:
Instrument: Agilent 6890B
Inlet Temp: 140 C
Inlet Liner: 1.0mm ID straight liner
Split Ratio: 1:5
Linear Velocity: 35 cm/s= constant flow-2.2 ml/min
Column: Rxi-624 Sil MS 30m, 0.32 mm ID, 1.8 um
Oven Ramp: 40 C hold 20 min, 10 C/min to 240 C, hold 240 C 20 min
Detector: 250C FID
Make up flow: 30 mL/min
Hydrogen flow: 40 mL/min
Air flow: 400 mL/min
Posted: Wed May 13, 2015 7:09 am
by Peter Apps
Two other things that we need to know about the setup:
Volume of headspace in the syringe, injection speed. ?
And you have not answered Bigbear's question about the rsd.
"The more you don't tell us, the more we can't help you."
Posted: Wed May 13, 2015 7:51 am
by ChemStation1
Injection volume-1ml
injection speed-300ul/s
RSD- about 10%
In every sequence, in 1 or 2 injections (from an overall of 6) the area of tert-butanol decreases by about 25%
Posted: Wed May 13, 2015 8:23 am
by Peter Apps
Your injection speed is 18 ml/min. This might be enough to disrupt inlet flow and split ratio. If you LOQ allows it you can decrease injection volume, keeping injection time the same, or increase split ratio. A wider inlet liner might help.
If you are consistently getting 1 or two results out of 6 that are 25% low, and the other four or five are in close agreement then I would suspect leaky vials. Check your sealing technique by putting a drop of pentane (or other low boiler) into a vial, crimp as usual, then immerse the vial in hot water, a stream of bubbles from around the cap shows a leak. Since you have an intermittent problem you need to do this for at least six vials.
Check whether it is consistently the same vials in the sequence that give low results. If it is you have a cool well in your oven (the thin aluminium shell on the CTC and clone headspacers does not spread the heat very well).