				Methylene Chloride interference at 8.4 minutes occurs only on one GC (see image link at end of message). When I switch the columns and inlet liner to another system (6890N) the interference at 8.4 minutes is gone. What could be causing this peak to appear on one system and not on the other?
Chromatographic Conditions:
System: Agilent 7890A
Columns: Stabilwax-DA 30m x 0.32mm x 1.0µm in tandem with Rtx-5 15m x 0.32mm x 1.5µm
Flow: 2.5 mL/min for 28 min \ 7.5 mL/min² to 10 mL/min hold 10 min (Helium)
Oven Temp: 35°C for 2 min \ 25°C/min to 60°C hold 7 min \ 15°C/min to 130°C hold 3 min \ 5°C/min to 180°C hold 0 min \ 30°C/min to 250°C hold 10 min
Splitless Inj: 1µL inj volume \ 50 mL/min purge flow \ 0.25 min purge time \ 4mm cyclo dbl gooseneck inlet liner (Restek Cat# 20895-214.1)
Inj / Det: 120°C / 250°C (30 mL/min Hydrogen 300 mL/min Air 27 mL/min Nitrogen make-up)
http://i.imgur.com/mbmwc41.png