I have looked up this topic on this forum and have taken some steps to try an resolve the issue. I am still having trouble. I am not seeing any peaks when injecting with or without a column. We were recently experiencing some issues with retention time of our Vitamin D3 peak. Our service engineer came out and changed the GPV, flow cell gasket, circuit board, and a few other consumbles. He checked the connections to the detector and they are ok. I am thinking there has to be air in the system, possibly the flow cell. All I get it noise from the mobile phase.
I have checked the injector volume and it is injecting the correct amount. The lamp is still good and all of the diagnostic tests are checking out. I have tried to flush the system with water, followed by MeOH. Any other suggestions of what may be causing this? Thank you in advance!