I am using a Shimadzu LC with a SIL-10AF auto sampler, for ethanol. I am currently experience an injection problem that I cannot figure out. The areas on my peaks are all very high, which is causing, obviously my concentrations to be too high. I have changed out the rotor, stator, needle and clipped tubing going into the needle. I also injected with my column disconnected 200ul of 70% methanol, to see if I had a plug that I just didn't find, I did this 12 times, then ran my mobile phase at the same volume, reconnected everything ran an injection of mobile phase and then my standard, that should be 12% ethanol, and it came up to 21%, does anyone have any suggestions what could be going on here?
I missed something here. After you disconnected, flushed, and reconnected everything, did you re-run the calibration? Also, just for completeness, what kind of column and mobile phase are you running, and was the retention time of ethanol consistent between when you ran your calibration and when you injected the check standard?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Found some plugging in a line that was apparently consistently moving until today! Thanks though
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry