Page 1 of 1

HPLC to UPLC

Posted: Fri May 01, 2015 1:15 pm
by gtma
What validation parameters should be evaluated to convert a validated HPLC method to UPLC? Should it be considered a "verification"?

Re: HPLC to UPLC

Posted: Tue May 12, 2015 1:02 pm
by artsjeroen
That depends on how much you've altered the method.

I have a method where I changed the column by a smaller particle column and decreased injection volume. For that method I have performed a verification (i.e., linearity / repeatability / recovery).

If the stationary phase (type), mobile phase, sample concentration etc. is altered, then I would perform a complete validation. But I guess it differ from lab to lab when to perform a complete validation.