Low plate counts of analyte peaks for EP monogrpah
Posted: Thu Apr 30, 2015 6:13 am
Hi all,
Low plate counts of analyte peaks was observed when verifying the determination of monomers (methyl methacrylate and methacrylic acid) in methacrylic acid - methyl methacrylate copolymer, a EP8.0 method (No.1127).
HPLC system: Waters 2695
Column: phenomenex nucleosil C18, 100mm * 4mm, 5um (EP recommended), brand new use.
Flow rate: 2.0 mL/min (EP uses 2.5 mL/min)
Wavelength: 202 nm
Mobile phase: methanol: pH2.0 phosphate buffer (30:70 v/v)
Injection: 50uL
Concentration: 0.26 ug/mL
Blank solution: 50 mL of methanol + 25 ml of mobile phase
The example chromatogram is showed as below:
http://imgur.com/zZmIO9r
The first two peaks are from blank; the last two peaks are analytes (methyl methacrylate and methacrylic acid). It shows low plate count for two analytes.
Additionally, the first analytes (1.557) was interfered with the peak from blank.
We try to:
1. Reduce the interference from the blank:
Action: We have changed different brand methanol but the peaks from blank were not reduced.
2. Adjust the retention time of analytes:
Action: flow rate was changed to 2.0 ml/min as flow rate at 2.5 ml/min per EP requirement showed low k’.
Please provide any advice for further investigation, thank you in advance.
Low plate counts of analyte peaks was observed when verifying the determination of monomers (methyl methacrylate and methacrylic acid) in methacrylic acid - methyl methacrylate copolymer, a EP8.0 method (No.1127).
HPLC system: Waters 2695
Column: phenomenex nucleosil C18, 100mm * 4mm, 5um (EP recommended), brand new use.
Flow rate: 2.0 mL/min (EP uses 2.5 mL/min)
Wavelength: 202 nm
Mobile phase: methanol: pH2.0 phosphate buffer (30:70 v/v)
Injection: 50uL
Concentration: 0.26 ug/mL
Blank solution: 50 mL of methanol + 25 ml of mobile phase
The example chromatogram is showed as below:
http://imgur.com/zZmIO9r
The first two peaks are from blank; the last two peaks are analytes (methyl methacrylate and methacrylic acid). It shows low plate count for two analytes.
Additionally, the first analytes (1.557) was interfered with the peak from blank.
We try to:
1. Reduce the interference from the blank:
Action: We have changed different brand methanol but the peaks from blank were not reduced.
2. Adjust the retention time of analytes:
Action: flow rate was changed to 2.0 ml/min as flow rate at 2.5 ml/min per EP requirement showed low k’.
Please provide any advice for further investigation, thank you in advance.