How to measure the I.D. of a capillary tube?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
7 posts
Page 1 of 1
I have some 1/16" OD stainless steel capillaries handy which are meant to be used as mobile phase pre-heat tubes in high temperature liquid chromatography. A trivial problem is that the internal diameters (i.d.) of the capillaries are unknown. Could anyone give me some advice how to measure them conveniently? Thanks in advance for any info.
-
- Posts: 421
- Joined: Wed Jul 18, 2012 6:45 am
Take two capillaries of known diameter and the same lenght, use them as loop in a manual injector valve. Inject a well absorbing substance (0.1 % acetone in water) and overfill the loops, then inject into a UV detector at 254 nm at low flow rate (0.2 mL/min or so, depending on the length of the capillary). Measure the duration of the acetone peak of the known diameters and compare with the unknown one.
Or, if you have a really good analytical balance, you can weigh the empty capillary, then fill it with water, weigh again and calculate mg = µL.
Or, if you have a really good analytical balance, you can weigh the empty capillary, then fill it with water, weigh again and calculate mg = µL.
-
- Posts: 327
- Joined: Sun Jan 01, 2012 10:21 am
Most capillaries have standard IDs such as 0.12mm , 0.17mm , 0.25mm , 0.5mm and 1.0mm.
You may even compare some of them ( 0.25mm , 0.5mm and 1.0mm ) by naked eye .
Use a magnifying glass for smaller IDs.
You may even compare some of them ( 0.25mm , 0.5mm and 1.0mm ) by naked eye .
Use a magnifying glass for smaller IDs.
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
The easiest is to put them under a stereo microscope with a measuring graticule.
Slightly more work; weigh them, fill them with water (or better mercury), weigh them again, calculate the mass of liquid, from that and the density calculate the internal volume, from that and the length calculate the cross sectional area and from that calculate the diameter.
Peter
Slightly more work; weigh them, fill them with water (or better mercury), weigh them again, calculate the mass of liquid, from that and the density calculate the internal volume, from that and the length calculate the cross sectional area and from that calculate the diameter.
Peter
Peter Apps
-
- Posts: 86
- Joined: Sun Oct 05, 2008 8:05 am
(Very) small modification of bunnahabhain's advice: you don't have to use them as injection loop, only connect unknown capillary to empty capillary with known ID. Inject (acetone) and repeat it again with only capillary with known ID. You might find volume of uknown capillary from elution volumes differences and then calculate ID.
HPLC 2017 in Prague, http://hplc2017-prague.org/
-
- Posts: 199
- Joined: Thu Jun 23, 2005 7:53 am
Why don't using a cheap (~20 EUR) USB-microscope to determine the inner diameter. Is the inner diameter really that inhomogeneous for steel capillaries?
-
- Posts: 16
- Joined: Fri Apr 24, 2015 2:44 am
Thank you so much for those good tips!
7 posts
Page 1 of 1
Who is online
In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 32 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 32 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science