Jaehwan Lee,
I would be surprised if you actually have an 11.7 lamp. That is not the most common since it is relatively short lived due to solarization of the window. For that reason it is suggested to keep it off when not in use and run it at the lowest possible temperature. To avoid the issue of the window you have to go to a windowless discharge detector like the HID or AID.
In addition, won't do water or methane although it will sometimes respond to some strange components like dichloromethane.
To do CO2 or methane you can look to either a TCD, HID or FID with a methanizer. To do water you are down to either a TCD or an HID (have fun with that one!)
Best regards,
AICMM