We're having an issue here in our lab running an assay of L-Glutamic Acid on our agilent 1100 hplc. We're doing a gradient elution with mobile phases of sodium phosphate 0.033 molar with a pH of 7.8(mobile phase A) and an acetonitrile 45% methanol 45% and water 10% solution(mobile phase B).
Our gradient program:

We're just using a Zorbax eclipse AA column with dimensions of 4.6 x 150 mm packed with 5 micrometer beadlets.
The issue we're running into is that randomly through our a run one or more of the injections have a steep drop in pressure that's unexplained that recovers in about 30 seconds. It's throwing off our assay. We're trying to nail down a cause but we haven't had luck yet.
Here is a image of the ideal pressure curve and then two pressure curves that had unexplained drops:
Normal

Abnormal:


As you can see the pressure drop isn't consistent as to where occurs in the injection and it happens randomly to one or more of the injections. The first time it happened everything was running smoothly for the system suitability and then when we finally got to our sample it happened. A bit frustrating considering each injection takes over thirty minutes.
The column has been in use for about a year and we've probably done close to fifty runs on it, we cleaned out the pump and we're really drawing blanks on what the problem could be.
We were hoping to not have to replace the column, but we will if necessary. We just replaced the guard column after having peak splitting issues.
Additional information can be provided at request.
Thanks you very much for any advice provided!