A have a 6890 agilent GC with 5973A MS.
I need to measure PAH with it but i have problem with calibrations.
We are using He as carrier gas and RTX-5SIL-MS ( 30*250*0,25)
The parameters are:
GC parameters:
Injector temp: 320°C, Splitless injection
Pressure: 8,73 PSI, Pulse pressure 36,3 PSI Pulse time: 1 min, Purge time: 1 min
Oven temp: 50°C for 1,5 min, then 20°C/min to 270, 5°C/min to 305°C, 40°C/min to 325°C and hold for 2 min
Column flow: 1,1 ml/min
MS parameters:
auotune
i have 10 SIM groups, MS source: 230°C and MS quad: 150 °C
I buy 2000 ug/ml PAH mix standard (contains 18 PAH + i make individually benz(e) pirene in acetone). I dilute them in 50:50% hexane: DCM solution. My calibration points are 25 ng, 100 ng, 400 ng, 1500 ng, 3000ng and 6000 ng/vial. I use naphthalene d8, acenaphthene d10, phenanthrene d10, chrysene d12 and perylene d12 internal standard for calibration ( concentration 400ng /vial and i buy mix for it too). I dilute my internal standards in acetone. I make everything in brown vials and storage in refrigerator.
I have problem with heavy compounds calibration. My curve fit is average of response factor. I can easily make 2-3% RSD for 13-14 compounds. My for last 5 PAH compounds i am unable to achive acceptable results (sometimes 50-60% RSD). At beginings a made my calibrations only in hexane, and i had the worst results with it. When i tried out 50:50% hexane: DCM, i could make good calibration for heavy compounds too ( all under 10% expect indeno(1,2,3cd)pyrene and dibenzo(a,h)anthracene 20-22%). It was about 2 week ago, now i wanted to again try it, i made 50 ng and 100 ng calibration points. For the first 14 compounds the recovery was perfect but the last 5 was horrible. The worst was dibenzo(a,h) anthracene i got 4-5 ng/vial instead of 50 ng, and 30 ng instead of 100 ng with my last calibration.
I cant understand my recovery of internal standards too. In my last calibration i got about 4 million area for napthtalene D8 for every calibration point and 3,2 million for my perylene d12. Yesterday I autoned again before i measured and for naphtalene i got about 3 million for every measure ( i tried it out with 8 different points) while for perylene 600 000- to 1,3 million aeras.
If it was volatile compounds, i could understand this thing. I tried out already so much thing if anybody got any clue what else to do i would be really happy!
Thanks for advance!
Erdno