Seems like there is a "newish" symptom that I see with the MS. When the MS is turned off (and turbovac allowed to wind down for 30+ mins), the deactivation of the roughing pump and N2 gases afterwards seem normal at first. But after ~30 secs of turning the roughing pump off, some gurgling would resume and oil would slowly rise up into the vac line with vigorous bubbling at first. The vac line can be vented by removing the clamp and gasket that connects the vacuum tubing to the back of the MS. This phenomenon was only observed after the "Source Exhaust Pump BAD" issue first appeared.
I also noticed that there is significant positive pressure in the turboionspray when the MS is activated (whether for equilbration or tuning) and the event/error log in Analyst records two individual errors: Source Exhaust Pump Failure and Pressure Too High in Ion Spray Chamber. The pressure can be vented when the ion spray is removed several times (plug in and out). There should be a slight negative pressure in the ion spray to aid with desolvation and also to vent any undesired solvent that slipped by the curtain plate and yet I have a large positive pressure. The tubing connecting the roughing pump to the turbovac then to the space between curtain plate and Q1 is under vacuum when I checked. Could this be a sign that a relieve valve behind the curtain plate is not switching properly?
Clement
Oil backing up into the line from the vacuum roughing pump would mean that the back flow valve is bad in the pump. When the roughing pumps are working properly they should hold vacuum in the system after they have been turned off. Also if oil it rising up that line, you may have too much oil in the pump itself. Be sure the oil level is between the lines on the window of the pump, if it is difficult to see you can remove the window when the pump has been drained and clean the window by rubbing it off with a cloth.
The positive pressure in the source sounds like the exhaust pump is not allowing any of the gasses and waste to vent out the bottom of the source. The area behind the curtain plate should be slightly positive because you run the curtain gas so that it flow out the orifice to keep out the unwanted solvent while the charge on the inner cone draws the ions in against the outward flow of curtain gas. The vacuum line that enters behind the inner cone creates the vacuum where the ions enter, and the line to it should be under vacuum and the valve is closed unless the instrument is actually operating, it opens when you begin to scan I believe.