Agilent 1200 system pressure over 400 bars

Posted: Thu Apr 16, 2015 8:51 pm
by maldiniyan
Hello, everyone. I am asking possible causes for a problem of our Agilent 1200 HPLC equipped with a G1311A quaternary pump. This system works fine for over 3 years, then suddenly one day pressure gradually increased within one minute into over 400 bars, which is the upper pressure limit and the system automatically shuts down.

To troubleshot this problem, first, I checked the cloggage on solvent inline filters which are fine, then columns, lines by disconnecting them from downstream direction, but the pressure still increased at over 400 bars even I disconnected everything downstream from the pump head and let the purge valve open.

Does anyone experience this same issue before and can someone give suggestions what are the possible causes? Thanks a lot!

Re: Agilent 1200 system pressure over 400 bars

Posted: Thu Apr 16, 2015 9:17 pm
by M Farooq
Have you cleaned check valves on the pump heads? Most likely dying seal parts have clogged the outlet check valves. Consult the manual for dismantling the check valves. They can be cleaned by sonication.

Re: Agilent 1200 system pressure over 400 bars

Posted: Fri Apr 17, 2015 5:46 am
by kubowicz.tomasz
Hello

Have you checked/replaced PTFE frit in purge valve?

Regards

Tomasz Kubowicz

Re: Agilent 1200 system pressure over 400 bars

Posted: Fri Apr 17, 2015 6:45 pm
by uzman
The outlet tubing of pressure sensor clogged.