ICP-MS: the effect of spary chamber temperature changing

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi, all,

Now, i'm using a Agilent ICPMS 7700, and 7900 in our lab.
My questions are as follows;
Now our spray chamber is 2degC.
Agilent model can be available from -5degC to 20degC for spray chamber.

I got the highest sensitivity in -5degC using 50ppt sample(H2SO4 solution) compared to 2, 10, 20degC.

I wanna know why the lowest temperature(-5degC) showed highest sensitivity.
Is anybody know this mechanism to explain about data?

Best regards.
Li
Hi,

At lower temperatures you remove more water from the aerosol and less water gets into the plasma. This also reduces the formation of interfering oxides in the plasma and you get a more stable response over time.

J
Hi,

At lower temperatures you remove more water from the aerosol and less water gets into the plasma. This also reduces the formation of interfering oxides in the plasma and you get a more stable response over time.

J

Hi, J
Thanks for your warmly respond for my question.
Thank you very much again.
Best regards.
Li
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 2 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: annnuu, Google [Bot] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry