weird sporadic detector noise
Posted: Sun Apr 12, 2015 11:53 pm
On one of our purge and trap GC-MS systems we are seeing random noise peaks appear. By 'noise peak' I mean non-Gaussian TIC peaks made up of mass fragments of varying abundance (although all within 1000 counts of each other in any particular scan). They are non-centroidal so appear as e.g. 101.1, 101.8, 102.5, 102.7, etc. and there are peaks across the full scan range (e.g. between 30 and 250 m/z there are 220 peaks). The peaks affect a cluster of chromatograms in the sequence and appear approximately every 2 weeks. The peaks do not appear at any particular retention time, they are randomly dispersed throughout the TIC. I'll post a pic if I can find somewhere to host it.
This first happened after ion source maintenance. We assumed the EM or HED had been disturbed so we checked the connection and blew away any dust with an air spray. we didn't see the noise until a few weeks later and it has periodically appeared since then. I assume the detector is recording a signal intermittently within a run, but I don't know if it is an electronics issue, or if something is physically creating a signal.
Has anyone else had similar issues?
System is an Agilent 6890/5973 MSD.