Thanks for the quick reply, Tom. Our current method uses a diverter to send the LC flow to waste both before and after our compound of interest has eluted, which has a RT of ~5 minutes. So it sounds like we should be alright.Actually, ion suppression should not be an issue. Phosphate is unretained in reversed phase; so long as your analytes are reasonably away from t0, the major factor should be simply the buildup of crud on the source. If you're going to be running a lot of samples, you might consider a diverter valve between the column and the source to cut the area around t0 to waste. Of course, that raises other issues with respect to band spreading, baseline stability, etc.