HPLC chromotogram errors
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
Hello dear colleagues.I have problem with my Agilent HPLC 1290 and i hope you will help me.I have new prepared calibration standards and good calibration curve.When i analyze any sample with this calibrated method the retention time have replaced. We use same solvent, same ambient conditions and chromatographic water for standards and sample preparation. And the result of our testing I can not see the amount of compounds. It relate with the replacing of retention time. I built the calibration curve and sample analyse in same day. What I have to do?
-
- Posts: 715
- Joined: Wed Aug 04, 2010 2:45 pm
Maybe it is a matrix problem, a problem with all the other compounds in your sample. I would try to prepare standards together with the matrix to see if the standard chromatograms will also change.
Good luck.
Good luck.
Gerhard Kratz, Kratz_Gerhard@web.de
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
Hello
1.Sample and injection solvent can affect retention time - try to dissolve sample in mobile phase and see if problem exists.
2.How big is retention time drift? sometimes when you have overloaded peak (API) you can see slightly difference for other peaks.
3.Check your calibration setting - you have tolerance range to automatically identify your peak based on retention time. Perhaps is less than 5%.
Regards
Tomasz Kubowicz
1.Sample and injection solvent can affect retention time - try to dissolve sample in mobile phase and see if problem exists.
2.How big is retention time drift? sometimes when you have overloaded peak (API) you can see slightly difference for other peaks.
3.Check your calibration setting - you have tolerance range to automatically identify your peak based on retention time. Perhaps is less than 5%.
Regards
Tomasz Kubowicz
3 posts
Page 1 of 1
Who is online
In total there are 25 users online :: 2 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: John Guajardo, Semrush [Bot] and 23 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: John Guajardo, Semrush [Bot] and 23 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science