Page 1 of 1
Frequency of Instrument Calibration
Posted: Wed Apr 08, 2015 6:03 pm
by ajrapport
Hello - is there any specific (USP/ICH) guidance stipulating the frequency with which instruments need to be calibrated (i.e., bi-annually, annually)? Thank-you in advance for your response!
Re: Frequency of Instrument Calibration
Posted: Wed Apr 08, 2015 10:35 pm
by KM-USA
We do OTC pharmaceuticals under cGMP. Our SOP states once per year. So that's what we do.
Re: Frequency of Instrument Calibration
Posted: Thu Apr 09, 2015 1:27 am
by tom jupille
To be nit-picky, an instrument should be *calibrated* daily for each method that will be run. "Performance Qualification" (which is what I think you're referring to) is typically done annually (in conjunction with regular preventive maintenance) and whenever maintenance / service / repair has been done on the instrument. The idea is to verify that it's operating properly before it is put back into service.
Re: Frequency of Instrument Calibration
Posted: Thu Apr 09, 2015 12:05 pm
by Gerhard Kratz
I agree with Tom, so I did after repair of the pump or injection system.
Re: Frequency of Instrument Calibration
Posted: Thu Apr 09, 2015 3:14 pm
by KM-USA
We do OTC pharmaceuticals under cGMP. Our SOP states once per year. So that's what we do.
I need to add that all cGMP or GLP assays have full system suitability runs bracketing samples as well.
Re: Frequency of Instrument Calibration
Posted: Sat Apr 11, 2015 8:06 pm
by Alfred88
We use the term "verification" instead of "calibration" for complex instrument like HPLC.
For the frequency, some firms perform the verification two times a year, while others do it once a year. This is "industry practice," and it is not defined in a guidance document.