Empower LC Calibration curve
Posted: Fri Apr 03, 2015 8:16 pm
by shinn
Hi does anyone know how to use empower to pick up peak areas for the standards automatically?
right now, I copy the peak area to the linearity curve window, and worst part is, I can not copy and paste multiple peak areas. Once at a time.
I tried to set up the "inject std" in the sequence, but did not work.
Shinn
Re: Empower LC Calibration curve
Posted: Mon Apr 06, 2015 1:27 am
by mattmullaney
Hi Shinn,
In the Processing Method, you can assign a name for each peak that elutes in the subsequent chromatogram. After this step, in the Run Samples window, you can assign which vials are used for the different levels of standard you're using (or one, if you're doing single-level calibration). Also in the Run Samples window, there is a button (or, in one of the menus, can't recall which one at the moment) called Amounts. Here you can copy all of the peak names from the Processing Method that you set up previously.
Between these steps, you've assigned the peaks' identities, set which vials the standards solution(s) are in, and told the software from where to get the data to form calibration curves.
This also may be done after a sample sequence is complete from the Alter Sample menu in the Browse Project Window.
Please, see what you think, and thank you.