Advertisement

Column installation

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi..
I have to installed a new column to GC (Agilent 7890A). How much length of column I should insert to inlet section and detector side?

Thanks in advance.
Hello

Please see below:

http://www.chem.agilent.com/en-US/produ ... olumn.aspx

Regards

Tomasz Kubowicz
2 posts Page 1 of 1

Who is online

In total there are 58 users online :: 2 registered, 0 hidden and 56 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: BB65, Google [Bot] and 56 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry