Very annoying background noise

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi all,
I have started using LCMS-QTOF Brukerfor three months and we have a problem on the instrument background in positive mode. there are two masses of 86 and 114 at 10*4 magnitude that are persistently present in the system using a HILIC column. We use ACN in B and water(MS grade) in C with formic acid 1% on both solutions and also 10mM Ammonium formate in solvent A. I have been washing the entire system with those solutions but the problem still persists. We also use a ramping approach for the solvents A and B.
Do you have any comments and ideas of how to get rid of these stubborn compounds?
Thanks for your help
It's not so bad (10e4). The background ions in the low mass range (<300m/z) is normal for ESI. If they don't interfere (interference with analyte ions, ion suppression etc.) - it's not problem :)

Look this:

http://sea.rice.edu/sites/sea.rice.edu/ ... dducts.xls
http://www.cigs.unimo.it/CigsDownloads/ ... 20Trap.doc
2 posts Page 1 of 1

Who is online

In total there are 29 users online :: 0 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry