Autosampler contamination?
Discussions about GC and other "gas phase" separation techniques.
4 posts
Page 1 of 1
Running VOCs (EPA524), and getting diminished response for 1,4-Dichlorobenzene-d4 and 4-BFB w/ respect to the other internal standards and surrogates. Area counts for these two standards slowly rise in successive levels of the calibration curve. There is also excessive carryover that needs multiple washes to get rid of. These issues do not occur when manually adding a sample fortified w/ internal standard/ surrogate directly to sparge vessel. Thinking this probably indicates contamination or active site somewhere in the pathway of the autosampler (Archon)? Flushing the autosampler with methanol, sonicating the equilibrium blocks, and replacing the water transfer line have not helped. Any suggestions to remedy this situation would be appreciated.
-
- Posts: 350
- Joined: Tue Jun 28, 2011 7:30 pm
I would wonder about the Archon syringe, the associated valve and seals on top of it, and the 6 port valve that adds the internals - this can get rusty over time. It may be worth replacing the line that runs from the Archon to the sparge vessel.
-
- Posts: 880
- Joined: Fri Jan 04, 2008 3:03 pm
Not familiar with that sampler. I just solved a similar problem on my Aquatek 70. When I modified the method so that there was no internal sample added, I got one of the 3 compounds I use. When I repeated this test with only MeOH in the IS vial, I got very much less of my FlBz. Come to find out the one way valve (check valve) suppling pressure to the IS vial was leaking. Replaced check valve and my problem is gone.
-
- Posts: 8
- Joined: Tue Apr 14, 2015 2:11 pm
Thanks to you both for your suggestions. I've already replaced the water transfer line with no benefit, so that is ruled out. With regard to the Aquatek70, I'm assuming the pressure check valve is analogous to the Archon's 6-port valve. I remember checking that for leaks without finding any, but with will have another look.
4 posts
Page 1 of 1
Who is online
In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 10 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science