- Posts: 14
- Joined: Thu Jun 29, 2006 4:22 am
I have a fluorescence issue that has me stumped and I hope someone on the forum can enlighten me.
Firstly, the setup:
The system is Dionex 3000 system with a Waters 2475 FLD (inherited from another lab).
I am looking at amino acids (glutamate and GABA) using precolumn derivatization with OPA + mercaptoethanol (all done in the autosampler) and run an isocratic MP (25% ACN, 1% THF, and 74% 50mM Na acetate buffer at pH4).
Detection settings: Ex = 337 nm, Em = 454 nm.
I use an internal standard that is added to the sample (by the autosampler) prior to the derivatization procedure.
A single run takes about 25 min and I run about 40-60 samples at a time with standards at the start, end and in the middle for the longer runs. My problem is that during the run something changes in the response of my samples (standards and unknowns). Basically, my response can be one of two types. 1) a low response from all analytes and 2) a much larger response so that my signals are close to maxing out (at a gain of 1). This wouldn't necessarily be a problem but the change in response is analyte dependent. For example, compared to the internal standard response the relative response of other analytes can be the same, lower or higher. Obviously this messes up any quantification.
I realize that your first thought is reproducibility, however, I have consistent responses for 25 samples in a row (response 1) and then all of sudden the response changes and the next 20-30 samples are consistent (response 2). It is like a switch has been flicked.
My thought is that it must be the detector because a change in wavelengths could alter the total response and the sensitivity to each analyte. Could it be that the detector is flicking between two different Ex/Em settings? how could this happen?
I have an example picture of two comparison standards but I am not sure how to attach it. I will happily email it to anyone if it would help or attach it if someone fills me in on how...
I hope I have explained this sufficiently.
Thanks,
James