H-P 5972 issues plus a question on PAH metabolite analysis
Posted: Mon Mar 23, 2015 7:26 pm
I have an old (I guess that's by definition) 5972, and a 6890 GC.
A few questions:
1. From longtime users of this instrument, what do you consider to be the high end of the concentration range you feel comfortable working in (say in ng of analyte put on the column, and in peak intensity, in scan mode)?
2. I find that sometimes after I run the autotune for standard spectra, the background goes up, sometimes to over 10,000. Clearly it's PFTBA. I've thought it seems like a faulty calibration valve, but it seems to happen sometimes and not others.
3. I have been asked to run some samples which are extracts from bacterial cultures growing on certain PAHs as sole carbon source. They want to quantify the parent compound and try to identify metabolites. The problem is that the concentration of the parent is huge (hundreds of ppm), but the metabolites are much lower. I can dilute them down and use a high split injection ratio, but then I don't think I'll be able to characterize any of the metabolites. I'm wondering if it would work to program the MS to turn off during the time the big parent peak comes off the column (like a solvent delay). These PAHs have 3-4 rings (so 175-200 MW), and the extracts are reasonably clean.
Thanks!
Harry
A few questions:
1. From longtime users of this instrument, what do you consider to be the high end of the concentration range you feel comfortable working in (say in ng of analyte put on the column, and in peak intensity, in scan mode)?
2. I find that sometimes after I run the autotune for standard spectra, the background goes up, sometimes to over 10,000. Clearly it's PFTBA. I've thought it seems like a faulty calibration valve, but it seems to happen sometimes and not others.
3. I have been asked to run some samples which are extracts from bacterial cultures growing on certain PAHs as sole carbon source. They want to quantify the parent compound and try to identify metabolites. The problem is that the concentration of the parent is huge (hundreds of ppm), but the metabolites are much lower. I can dilute them down and use a high split injection ratio, but then I don't think I'll be able to characterize any of the metabolites. I'm wondering if it would work to program the MS to turn off during the time the big parent peak comes off the column (like a solvent delay). These PAHs have 3-4 rings (so 175-200 MW), and the extracts are reasonably clean.
Thanks!
Harry