Hello to all,
I'm trying to develop a method for impurities determination. I'm using a mobile phase with 50% ACN - 50% H20 and a solvent with 30% ACN. The injection volume is high, 300 μl because the low concentration of my product. When i run in HPLC the ref solution of 1% (dilution 1ml of test solution to 100 ml with solvent) then an appropriate peak displays in my chromatogram. When i run a spiked sol 1% (dilution of 1 ml of test solution to 100 ml with placebo) then the peak of interest is poor with very low area. I noticed that the baseline with placebo is higher so the baseline covers my peak. If i use lower injection volume then the peak improve but i cant use lower injection volume because of my products low concentration. Can you help me how can i fix my baseline without modify the injection volume? Or if there is any other solution i can try?
Thnx in advance
I noticed that the baseline with placebo is higher so the baseline covers my peak.
Does that mean that your analyte peak is lost in the tail of a placebo peak? If you could post a link to a chromatograms (analyte only, placebo only, and spiked placebo), it would be easier to advise. Instructions for linking are here: viewtopic.php?f=1&t=2617

If it is a case of mass (of the placebo) overload, then changing the injection volume won't help; what you would need to do is to improve the resolution between the interfering placebo peak and your analyte.

If it is a case of volume overload, you may be able to improve things by using a weaker diluent (100% water if possible). Your current diluent is already weaker than your mobile phase, which is a good thing, so I don't know how much additional improvement is there. All you can do is try.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thank you for the fast response Tom. Well today i tried lower injection volume and i got the ideal chrom with 50μl inj volume. The area of the peak is far too low so i must try something else. My products formula includes high amounts of NaCl so i think thats the reason why at 2mins appears a huge peak in my chrom (overloading the column) which affects my analyte at 9mins. I'm thinking of trying to start my method with 80-20 H2O-ACN so get rid of the salt and then change to 50-50 for elute my analyte. What do you think of this idea?
Thank you again
I'm thinking of trying to start my method with 80-20 H2O-ACN so get rid of the salt and then change to 50-50 for elute my analyte.
I generally try to avoid "step" changes (differences in mixing characteristics can make them difficult to transfer), but in this case I think it is worth a try. The other thing to try might be to use less ACN in the diluent and see if that lets you increase the volume again.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
We'll I tried changes in solvent,even 100% water, with no improvement. I have good results using as mobile phase 20% ACN for at least 5mins and change to 50% ACN. So now I just have to make some small changes to optimize the gradient to display the rest impurities peaks in an acceptable time.
Glad to hear it's working out! :D
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
6 posts Page 1 of 1

Who is online

In total there are 28 users online :: 2 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Baidu [Spider] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry