precision problem

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
hi sir,

iam facing problem with GC calibration.

The details are
Gas chromatograph; Make : Shimadzu; Model GC-2010
Standard ; 1250ppm of biphenyl in n-heptane, 5 injections in single vial
Capillary column; BP-1 30m,0.53mmID,1µm
Injection Volume;2µl
Column oven Temp; 130°C
Injector Temp; 250°C
Detector Temp;250°C
Split Ratio;1:1
Viscosity delay time ;0.2sec
Column flow;10 ml/min
total flow; 23 ml/min

And finally we get the %RSD is 3.9 against 2.0%.

please find What problem going with these parameters.

Thanking you sir
Sreenivas.B
Divi's laboratories ltd
Vizag
Mail;bobburi496@gmail.com
A split ratio of 1:1 combines all the disadvantages of splitless and split injections. Either go for splitless with a splitless tiem of about 30 s and a purgefl owo f 30 ml/min, or go with a split injection - your concentration is plenty high enough to use split. Also a 2 ul injection gives a bigger vapour cloud than 1 ul, which can also cause problems. If you need splitless or a 1:1 split and a 2 ul injection to get big enough peaks than you probably have a problem with the detector.

Peter
Peter Apps
I'm not sure about Shimadzu but Agilent instruments cannot generate a <5:1 split reliably due to the extremely low gas flow to the inlet. That is likely the source of your problems. I agree with Peter, Go splitless with a low injection volume and short splitless time or go at least 5:1 split with a larger volume. I tried a pulsed split when doing headspace with a 3:1 split and was getting RSD's near 20% due to that issue so I went 5:1.
I agree with the above statements, and will add, you should consider reducing your injection volume. In our Agilent instruments the largest inlet liner can contain only 990ul of vapor.
4 posts Page 1 of 1

Who is online

In total there are 127 users online :: 1 registered, 0 hidden and 126 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Amazon [Bot] and 126 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry