Powerful flush of Rx-Sil column
Posted: Mon Mar 16, 2015 10:02 pm
Hi all,
I had a mishap with an rx-sil column today and am looking for a way to wash/restore the column. Not sure what the best strategy is for flushing everything out.
I am using a preparatory reverse-phase rx-sil column with ethanol and water as solvents (95-100% etOH). The sample I'm injecting is in DMSO due to solubility issues.
We always make the manual injection with a standard 3mL syringe. Today I used a 1mL syringe, of the exact same make and model. Apparently, however, this syringe features a different plastic and rubber stopper that is NOT compatible with DMSO. Unbeknownst to me, as I injected my sample, I was also injecting some dissolved plastic/rubber from the syringe. Column was not happy, pressure immediately jumped past 400 bar.
I have replaced the guard column and washed the column on reverse for 2 hours with 70/30 Etoh/water. Replacing guard column had little effect on pressure. On reverse flow, pressures were normal for our method (~230 bar). When I return to forward flow, pressures remain high (~350 bar). I tried cleaning the inlet frit as well as possible without removing it. There aren't any visible blockages. Also confirmed there aren't blockages elsewhere in the system.
I'm currently reverse washing with 50/50 EtOH/water. Wondering if anyone has ideas for a powerful but compatible wash solution to really get rid of all this plastic.
Let me know your thoughts/ideas.
I had a mishap with an rx-sil column today and am looking for a way to wash/restore the column. Not sure what the best strategy is for flushing everything out.
I am using a preparatory reverse-phase rx-sil column with ethanol and water as solvents (95-100% etOH). The sample I'm injecting is in DMSO due to solubility issues.
We always make the manual injection with a standard 3mL syringe. Today I used a 1mL syringe, of the exact same make and model. Apparently, however, this syringe features a different plastic and rubber stopper that is NOT compatible with DMSO. Unbeknownst to me, as I injected my sample, I was also injecting some dissolved plastic/rubber from the syringe. Column was not happy, pressure immediately jumped past 400 bar.
I have replaced the guard column and washed the column on reverse for 2 hours with 70/30 Etoh/water. Replacing guard column had little effect on pressure. On reverse flow, pressures were normal for our method (~230 bar). When I return to forward flow, pressures remain high (~350 bar). I tried cleaning the inlet frit as well as possible without removing it. There aren't any visible blockages. Also confirmed there aren't blockages elsewhere in the system.
I'm currently reverse washing with 50/50 EtOH/water. Wondering if anyone has ideas for a powerful but compatible wash solution to really get rid of all this plastic.
Let me know your thoughts/ideas.