methanol and permanent gases
Posted: Wed Mar 11, 2015 3:28 pm
Hello,
I have a Perkin Elmer Autosystem XL with a 60/80 carboxen 1000 (15 ft) column and TCD detector. I currently have excellent results measuring CO, CO2 and CH4. Now, I need to measure methanol and I cannot detect it. I don't see the methanol exiting the column even when I let the program run for 40 minutes. The procedure I am using is the following: Carrier gas He (30 ml/min). 35°C (5 min) to 225°C at 20°C/min.
Is it expected that methanol would exit at very long times?
Do you think another column is more suitable for the separation? I have a 60/80 Porapak Q (6 ft), 100/120 Hayesep D (7 ft) and a MolSieve 5A (6 ft). Two of them in series could be a solution?
Any advice will be greatly appreciated.
I have a Perkin Elmer Autosystem XL with a 60/80 carboxen 1000 (15 ft) column and TCD detector. I currently have excellent results measuring CO, CO2 and CH4. Now, I need to measure methanol and I cannot detect it. I don't see the methanol exiting the column even when I let the program run for 40 minutes. The procedure I am using is the following: Carrier gas He (30 ml/min). 35°C (5 min) to 225°C at 20°C/min.