poor sensitivity of ketonoes using Archon in the soil mode

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hello:

We have been using our Archon autosamplers in the water mode and ketones look good. We started calibrating the VOCs in the soil mode and aside from ketones all the other compounds look good. Acetone and MEK have very low responses and we have not been able to get any calibration for these two compounds! We are preheating the container to 40C and using 5ml sample amount. Any suggestions?

Thank you very much.
It could be an active site in the transfer line or soil probe. Compare the recovery of a high point and low point purged in water mode to soil mode. If the high points are similar but the low point has a low response in soil mode it suggest an active site.

If that is the case try backflushing the transfer line or replacing it.
I would also wonder if the soil transfer line and valve heaters are on and working.
3 posts Page 1 of 1

Who is online

In total there are 28 users online :: 0 registered, 0 hidden and 28 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 28 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry