Page 1 of 1
Accela LC PUmp: not running, lights flashing
Posted: Tue Mar 10, 2015 2:26 pm
by Arne
Hi All,
time to share one of my problems again. :|
We have been using a Thermo Accela LC Pump in our lab with great success but a few days ago it suddenly stopped working and gave only a generic error message. Restarting the pump and computer and checking the fuses does not solve the problem.
The Power and Communication LEDs are flashing green at a different rhythm than usual. The LEDs are off for about 1s and then flash for about 0.1s. Sometimes the communication LED flashes orange.
Has anyone seen this problem and maybe even knows what's wrong?
Thank you so very much for your advice,
Arne
Re: Accela LC PUmp: not running, lights flashing
Posted: Wed Mar 11, 2015 2:34 am
by tom jupille
The best advice I can give is the obvious: get hold of your Thermo rep. He/she should be able to.point you in the right direction.
Posted: Mon Mar 16, 2015 3:35 pm
by Arne
Tom,
thanks. We did figure out the problem: The active dynamic mixer had produced a short circuit. We replaced that part and the pump is back in action.
Arne
Posted: Mon Mar 16, 2015 11:36 pm
by tom jupille
Thanks for the update (it may save the next person some grief!).