Hi, we have a prostar 320 detector and it showed a message "No light at current λ".

So I change the lamp but now it shows "No Light - Calibration Error".

Any suggestion? Thanks
Possibilities:
- the new lamp is also dead
- the new lamp is not properly connected or aligned
- the power supply to the lamp is not functioning
- something is blocking the optical path

Is there any evidence that the light is, in fact, lit (sound? a faint whiff of ozone? )?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Possibilities:
- the new lamp is also dead
- the new lamp is not properly connected or aligned
- the power supply to the lamp is not functioning
- something is blocking the optical path

Is there any evidence that the light is, in fact, lit (sound? a faint whiff of ozone? )?

The lamp is okay because I can see the light and hear the sound.

I think there's probably some thing blocking the optical path.
3 posts Page 1 of 1

Who is online

In total there are 23 users online :: 3 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Bing [Bot], Semrush [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry